找回密码
 马上注册

扫一扫,访问微社区

搜索
热搜: mestrenova
查看: 3269|回复: 2

自动匀场

[复制链接]
发表于 2006-11-16 21:49:43 | 显示全部楼层 |阅读模式

Autoshimming

Interactive autoshimming should not be used as a replacement for manual shimming. Autoshimming is fast and efficient only when the shims are near optimum. Autoshimming only on Z1 and Z2 is recommended. Do not autoshim on shims other than Z1, Z2, and Z3.

1. In the vnmr acquisition shim window, click on auto.

2. Select m>m to "start with medium and end with medium" shimming quality.

3. Select the set of shims to be used by the autoshimming routine.

4. Click on start. The start button will change to stop. The autoshim is finished when the stop button changes to start.

Use background autoshimming for long data acquisitions EXCEPT for 15N and spin-lock experiments:

1. While setting up your experiment parameters, type method('z1z2') for samples > 0.5 ml, type method('z1z2z3') for samples £ 0.5 ml.

2. After the experiment has started, type shim.

Table 1 lists the room temperature shims. If you change a shim, you should reoptimize the other shims which interacts with the changed shim. Shims which start with 'Z' (the spinning shims) should be optimized with the spinner ON, while shims that start with 'X' or 'Y' (the non-spinning shims) should be optimized with the spinner OFF.


Shims (ethylbenzene in chloroform): Linewidth = ______ sidebands = _____

Shim

Interacts With

Value

Shim

Interacts With

Value

Lock

none


X1

none


Z1C

none


Y1

none


Z2C

Z1


XZ

X1


Z1

none


YZ

Y1


Z2

Z1


XY

X1,Y1


Z3

Z1,Z2


X2Y2

X1,Y1


Z4

Z1,Z2


X3

X1,Y1





Y3

X1,Y1

% G: i- g% z0 [: {8 y% {2 [4 e" `
[此贴子已经被作者于2006-11-17 19:08:38编辑过]
8 x( u* ~/ ]9 Z% R" p# \6 y8 v
回复

使用道具 举报

发表于 2006-12-13 13:28:44 | 显示全部楼层

sf

严重支持

回复 支持 反对

使用道具 举报

发表于 2008-1-6 09:52:58 | 显示全部楼层
xiexie
回复 支持 反对

使用道具 举报

您需要登录后才可以回帖 登录 | 马上注册

本版积分规则

Archiver|手机版|中国核磁共振论坛

GMT+8, 2024-11-26 20:45

Powered by Discuz! X3.5

© 2001-2024 Discuz! Team.

快速回复 返回顶部 返回列表