发表于 2010-7-18 23:49:41
|
显示全部楼层
|
1) atma 1H (Tunning+Maching)
5 G9 {; w5 T. j+ I p2) take a normal 1H spectrum with ns = 1; rga , etc
; x' Y6 t1 |; f! w3) ft and phase it manually or just type apk% D; k2 N2 g5 i9 i- H7 I
4) change pulprog zg30 to zg, using eda or select AcquPars icon
- u3 F: N) G7 f4 |2 p: _5) change the p1 value to (p1 plus or minus 0.1 or 0.2 and so on) *4
- @4 R5 A3 |' `; z6) check the p1 value, (it should be below 50u) ->type p1 return
! r6 ^$ a) |( X% p1 T% I7) zg again( T$ Q2 D/ \1 Z# m& l/ ~5 w+ e
8) fp (this moment do not phase the spectrum anyway): ]; w& _9 e# d6 r+ e0 r
9) if the peaks appear mostly positive, reduce p1 value, and do step 7 and step 8 again) v" p4 h+ r& D5 o
10) if the peaks appear mostly negative, increase p1 value, do step 7 and step 8 again" {% f6 Q7 l" ~0 V
11) while the intensity of the peaks appear mostly half positive and half negative, then it is the correct 90 pulse width *47 }5 G4 Y- m& i+ t6 l$ m2 \
12) divide p1 value by 4 (this is the value which can be used in 2D expts for 1H 90 pulse hard pulse).
3 a9 s1 x7 Q! w13) change pulprog = zg30, ns = 16, p1 =(90pw) start the 1H experiment( Reset to Normal Mode ), K- J& n* z+ @8 z6 k$ R
Note :If no appropriate value is find at all, start the small value of p1=3u,with pulprog = zg again , then check that the phase is positive or not. It should be positive, else, try again from step 5 to step 12.. r) ]; ?9 B1 F1 m
|
|