1. lock field shifted? lock phase correct? what is lock power? is correct solvent always choosed when locking? 2. tuning/matching is OK? 3. spinning is on or not? what is airflow? 4. topshim1d failed for all the solvent or specific solvent? 5. swap SGUs in edasp and run topshim to see if there is any difference. I saw one of this kind of problem caused by bad SGU. This can be confirmed by any recent sensivity loss in experiment
it is easy to change setting in edlock, and confirm the saving with a password when exiting.
以下是引用lynnsu在2009-2-13 21:16:08的发言: 4 o# Y+ N4 d" I) D ; q2 r0 N4 r& B3 b) a# h, t( r% E
我的意思是如果机器空闲的时候不锁场会不会造成场漂呢?机器有时候不做样的时候长时间不锁场。
* f! U3 D) V4 u7 I2 H, F y! f. a; L; v; [% i- x+ X
major field shift only happen after new installation or certain hardware change. the minor shift, however, happen all the time. If the shift is still within the range which can be handled by lockshift. The autolock should take care of the lock for you, otherwise you have to use your CDCL3 sample to locate the lock signal right in the middle of lock window, and update the field value using edlock.